Hi Roldan,
This is in continuation to Howards reply.
FDA shall not come and audit only functioning of cleanroom.
However, in cleanroom qualification the main considerations are, not limiting to
1) AHU / HVAC design considerations and qualification
2) Non-viable particulate count. This needs to be online for ISO 4 / 5 / 6 area.
3) Viable particulate monitoring (refer USP for test details)
4) Cleaning of cleanroom.
5) Monitoring of Environment conditions (temp. & relative humidity). Differential pressure across the rooms.
6) Personnel qualification, entry exit procedure
7) Material inward outward flow
Material used for floor, ceiling, walls. Are the corners coved to avoid particulate accumulation.
9) How is the material handled / processed under cleanroom (under LAF or how)
10) Qualification of utility viz. Compressed air, Nitrogen, Water.
Shall advice you to have a risk assessment done for the cleanrooms considering the activities performed.
For eg. In one of my assignments of manufacturing empty pre-fillable syringes (glass), there were activities like needle bonding and then cleaning. Both were performed in ISO Class 6 (at rest). Still we didnt have online particle counter. The justification was, after needle bonding we had a washing & siliconization machine wihch was automatic without human intervention and activity performed under machine LAF.
This way we saved the budget and utilized in QC instruments.
Incase u still have queries, kindly elaborate your question.