You may need to take some professional guidance because you are getting into areas where lack of knowledge could expose you to people like the FDA who allow no deviations and take no prisoners in the manufacture of controlled materials like medical devices.
Basically ISO 7 is OK for things like catheters and stents but you also need to control the bioburden by taking swabs and using agar plates. Bacteria is the big issue and its controlled by the various Health bodies who usually refer to the "orange book", the US pharmacopea or the EU GMP guide to good manufacturing Practice (GMP) .
They have 4 types of room, Grade A,B,C and D and generally follow the 14644 ISO levels but they also insist on measuring 5 micron particles. 20 /cu mtre grade A , 29 B , 2900 C and 29,000 for D. with 0.5 micron rates of 3,520, 3520, 352,000 and 3,520,000 per cu metre. Therfore grade A needs continual particulate monitoring and grade B is usually turbulent so compressors can be used. However they would have to be oil free to a very fine level and all cylinder exhausts need to be piped out the room.
Theres a mass to do and without experience your man is heading for disaster
Regards,
Dick Gibbons (committee member of the BSI LB I 30 in the UK and convenes 14644 part 8 and 10 for Chemical contamination)
I agree with you that this scenario is very sensitive. So you will need to gather more pertinent information and guidelines in medical devices. Absolutely, it doesn't need an impromptu but requires planning to prevent from risks or threats.